Analysis codes
calibrations for bin_dgs in GEBSort_nogeb
GEBSort_nogeb is the program that can analyze data from DGS, DFMA and GRETINA.
You can get if from here:
cd workdir (e.g., /home/gtuser/gebsort) svn co https://svn.anl.gov/repos/gs_analysis/GEBSort . or wget http://www.phy.anl.gov/gretina/GEBSort/AAAtar.tgz tar -zxvf AAAtar.tgz
To produce the PZ spectra and 2D [sum2-sum1] vs [sum1] matrices needed to determine the PZ fudge factor (FF) you must enable
#define ALL2DS 1
in bin_dgs.c and recompile . We do not always want these spectra as they take up a lot of space, but for now we need them
You now specify the PZ and ecal files in the GEBSort.chat file with these lines:
dgs_MM 350 dgs_PZ dgs_pz.cal dgs_ecal dgs_ehi.cal
For DGS data, enable bin_dgs in the GEBSort.chat file. To find the PZ values to use, sort some data from a 207Bi source. Then extract the pz spectra in .spe format with the get_pz.cc script
GEBSort_nogeb ....
rootn.exe
dload("bi.root")
.x get_pz.cc
Now run:
dgs_pz 350 100 dgs_pz.cal 1.003
where 350 100 are the M and K values you find in the runxx.save file. Specify the values in 10 nsec units. In this case I saw these lines in the .save file:
caput GLBL:DIG:d_window 0.06 caput GLBL:DIG:k_window 0.20 caput GLBL:DIG:m_window 3.50 caput GLBL:DIG:k0_window 0.80 caput GLBL:DIG:d3_window 0.20 caput GLBL:DIG:raw_data_window 0.32 caput VME01:SDIG1:k0_window0 0.0 caput VME01:SDIG1:k0_window1 0.0 etc
for the K value: sum up all the K and D values, in this case: 0.06+0.20+0.80+0.20 = 1.26 us or 126 in 10 nsec units. Notice that what is considered the K value also includes the D values (per SZ 6/25/18) as well as a D2 which is fixed at 0.15 (per JTA 6/26/18) and not in the listing above because the user cannot set it. Thus, in total, K in this example is 1.41 us or 141 in 10 ns units. The M value is 3.50 us or 350 in 10 nsec units. The 1.003 is a modification factor that needs to be determined by looking at energy vs baseline spectra.
you already specified the M value in GEBSort.chat
Now, after the PZ file is generated, remove the energy calibration file if there is one:
rm dgs_ehi.cal
so that the calibrations defaults to 0 and 1 for offset and gain and sort again using the new pz values that were extracted above. When you resort, the PZ values in dgs_pz.cal are read in and used. Extract the new clean, uncalibrated, ehi spectra as
.x get_ecln.cc
and run the calibration program
dgs_ecal dgs_ehi.cal 207Bi 2000
you can also use "88Y", "60Co" for the source. The calibration will be 1keV/ch. The last parameter specifies the lowest channel the program will search for peaks in to avoid noise at low energies.
Next when you run GEBSort_nogeb, both the new PZ values in dgs_pz.cal and the gain and offset values in dgs_ehi.cal are read in and used.
The energy processing in bin_dgs follows algorithms that were developed by Shoufei Zhu.
