Detailed Steps for Tuning In-Flight Beams
Overview of the steps for In-flight beam delivery
Once steps for degraded beam tuning have been completed and a good focus is found at the RAISOR midplane (Separate detailed sheet for this), the steps below can be used to help guide the identification and tuning of the secondary beam.
- Acquire information on primary/degraded beam tunes and prep for beam
- Ask for docs from operations on primary and degraded beam TOFs and the charge state and slit scan data to be loaded into BOX
- Acquire the final cup/collimator FC currents and any notes on degraded beam transmission
- Note down RAISOR slit settings (in particular the midplane vertical slits are key)
- Insert detector of choice, make sure HV and preamps are on and daq is alive w/ noise etc
- Getting a reference spectrum from the degraded beam (can be done at RAISOR exit or target depending)
- Cut back beam by x100 w/ slits (not RAISOR slits) first then insert all available attenuation (~x15 Million) + RAISOR Att if needed
- Place beam on detectors and slowly remove attenuation until rates are roughly >100 Hz & <500 Hz
- Take ref spectrum - optimize daq params and save the settings file as they should not be changed for the duration of the run
- Run the RAISOR midplane slits in to 2 - 3 mm gap and take another spectrum (better to have small slit spacing for BRho scans)
- Determine the RF Sweeper(R501)/R401 phases for degraded beam with known TOF
- Set primary degraded beam rate to few hundred Hz and get Grafana running
- Have operations turn on Sweeper to Voltage=2 and phase=0 to start (R401 is still off)
- Note: You already may see a reduction in the beam rate do not increase beam current to get back to few hundred Hz
- Set the increment step = 1 degree, operators hold down increment to scan over 360 noting the phase every 10 - 20 degree
- Document the phase that transmission is the highest for the primary degraded beam
- Do smaller steps of ~1 degreed around the regions of the maxima (should be two that are similar one <180 one >180 deg)
- Leave Sweeper(R501) at optimal phase, increase amplitude to 3,4,5,6V to see if improvement in transmission, note the best value.
- Set Sweeper (R501) to optimal phase/amplitude
- FYI: Phase B is the number that ops has been typically used when referring to the phase
- Turn on R401 to phase = 0, amp ~ 1.1 or 1.2 (below 0.9 seems to give it issues locking up)
- First scan phase, then scan amplitude similar to Sweeper to find optimal R401 parameters, note the values.
- TURN OFF THE SWEEPER AND R401 FOR NOW (or maybe only needs to have V=0??)
- Should be back to the few hundred Hz rate as at the beginning
- Identify the secondary beam and optimal scale factor
- Have a good degraded beam spot in the detectors (SCALE = 1.0), w/ RF Sweeper and R401 OFF
- Best to have RAISOR slits set to 2 - 3 mm spacing only (typically they are 5 - 7 mm after RAISOR Proc.)
- This can be done either with low primary beam on FCA001 or higher beam and RAISOR x100 or x1000 in (just document)
- Input current beam line values into ATLAS RAISOR Scaler Tool (make sure it is for the appropriate beam line)
- Incrementally scale the RAISOR / beam line magnets in 0.5-1% steps (up or down depending) until 1% or so past expected scale value
- At each step, remove attenuation to achieve the 100-500 Hz rates and searching for the beam of interest
- Take a fixed time run with the daq to note rates / purities
- Generate cuts at each step to track the movement of the beam charge states etc. if needed
- Identify the isotope of interest based on calculated energies or previous production of the isotope
- Calculate the rate of the isotope of interest as the pps / pnA [note: pnA value from FCA001 divided by q of primary beam]
- Optimize the secondary beam
- Set beam of interest to optimal scale factor
- Put max current on FCA001 and RAISOR att in if needed (essentially as close to primary beam maximum on target as possible)
- Carry out procedure for optimizing the RF Sweeper and R401 as described above [Determine the RF Sweeper(R501)/R401 phases for degraded beam with known TOF]
- There are a few options at this step:
- The RAISOR midplane slits may be opened to 5 - 7 mm gap if the purity is good enough prior to the phase scans
- Otherwise, the slits may be opened after finding the optimal settings for the secondary beam to see if rates are increased
- Finally, if secondary beam rate is too low relative to the total rate for a manual scan, we can calculate the expected phase and search nearby this theoretical value
- Leaving the Sweeper / R401 settings at the optimal values, provide rates / purities etc. to operators for optimizing on the total rate
- Keep track of primary beam current at various points through tuning, RAISOR POC needs to help determine with discussions on ops/user sides when to stop tuning
- Once rates have been achieved, move to any other optimizations, i.e. beam spot etc.
- Document and Monitor
- Note down RAISOR settings when needed
- Keep track of gas-cell temp pressure when applicable
Definitions
- primary beam - the beam species and charge state taken out of the source at the ATLAS energy requested
- degraded beam - the same beam species as the primary beam but at a different (lower) energy and most likely different (higher) charge state
- production target - the target put in place before RAISOR, e.g., Be foil or Gas Cell